KM56c problem
« previous
next »
Pages:
1
[
2
]
All
Go Down
Print
Author
Topic: KM56c problem (Read 11601 times)
Kai
Sr. Member
Offline
Posts: 438
Re: KM56c problem
«
Reply #15 on:
December 30, 2016, 05:28:12 AM »
I would ask you to redo the measurements after making your self familiar with your meters:
First identify the pins, according to the schematic above.
Let's name the pins 1 to 3 from top to bottom.
Now you need to measure capacitance and resistance of each pair.
1-2 C= 45pF, R= >10GOhm
2-3 C= 45pF , R= >10GOhm
1-3 C= 1025pF (+/- 10%), R= >5GOhm
As your meters probably cannot measure higher than several hundred MOhm, simply check if they read "no conductance" in the highest resistance range.
If there is an Ohm reading value on any pin combination your capsule mount is faulty.
As this does not necessarily mean the capsules themselves are broken, you should consult a very skilled person to fix it.
I would strongly suggest not to mess around by yourself, as these precious diamonds are not repairable, it's very likely you finally break them if you try.
Logged
afterlifestudios
Full Member
Online
Posts: 200
Real Full Name: John Raham
Re: KM56c problem
«
Reply #16 on:
December 31, 2016, 06:13:56 PM »
Thank you, Kai. Here are the results of my capsule measurements according to your pin out nomenclature.
pin 1-2 C = 1079pf R = no conductance
pin 2-3 C = 0pf, R = 3.8 kOhm
pin 1-3 C = 1079 pf, R = no conductance
Clearly there is a problem. I would love to send it to a "very skilled person" to fix it. As far as I know from comments made earlier in this thread by Klaus, Thiersch only reskins diaphragms for these capsules. Based on the photos of my diaphragms and the readings I'm getting on the capsule pins, does it really seem like a diaphragm problem? If so, great, I'll send it to Thiersh. If not, does anyone have a recommendation for a "very skilled person" willing work on it?
Logged
uwe ret
Jr. Member
Offline
Posts: 81
Real Full Name: Uwe Sattler
Re: KM56c problem
«
Reply #17 on:
January 01, 2017, 02:36:12 PM »
It looks as if your capsule head has an internal problem between the back plate and the rear diaphragm. Kai was close with the expected measurement values. The actual values between either diaphragm and back plate should be 70 pF (35 pF in parallel with the series combination of 1000 pF and 35 pF), and the capacitance between the two diaphragms should be 1018 pF (1000 pF in parallel with the series capacity of the two 35 pF capsule capacitance). Resistance should be infinite for measurements between any of the 3 terminals. As is the capsule could still be used in cardioid only mode.
Logged
afterlifestudios
Full Member
Online
Posts: 200
Real Full Name: John Raham
Re: KM56c problem
«
Reply #18 on:
January 01, 2017, 09:46:51 PM »
Thank you, Ewe. Indeed, that's what it seems. But I don't have a healthy cardioid "as is". Perhaps if I completely lifted the rear diaphragm from the circuit, then the front side would be normal cardioid? (I'm not ACTUALLY going to do that!) But right now it definitely is cardioid only, but very low level. Handling or tapping mechanically on the mic housing gives very full, loud clear results. Just not from the capsule.
Could the fact that the rear side is shorted to the back plate be causing that problem with the front side?
John
Logged
afterlifestudios
Full Member
Online
Posts: 200
Real Full Name: John Raham
Re: KM56c problem
«
Reply #19 on:
April 07, 2017, 07:45:33 PM »
Just a happy follow up here...
Once it had been determined that the problem was capsule related, not amplifier, I sent the capsule to forum member "Panman". He found the short between capsule and backplate and corrected it.
I got the capsule back from him today and it sounds marvellous in all 3 patterns. Output level in cardioid is generally on par with my km84's. Noise floor is about 5db higher than the 84's.
Most importantly, it sounds great! I thought she was a goner...
Thank you, Panman!
John
Logged
Clomi
Newbie
Offline
Posts: 1
Real Full Name: clomifuge
Re: KM56c problem
«
Reply #20 on:
January 26, 2024, 10:46:21 AM »
Hi all happy new year,
Somebody have experience with the KM56-C calibration Input PIN ?
Best and thanks
Clomi
Logged
klaus
Moderator
Hero Member
Offline
Posts: 2227
Re: KM56c problem
«
Reply #21 on:
January 26, 2024, 01:04:09 PM »
'clomifuge' is a name? First name? Last name? Avatar name?
Please correct your post and insert your real, full name.
Thanks,
KH